Calibrating High Power Sensors

A community of technical professionals involved in the calibration of RF power Sensors.
Post Reply
csperrazza

Calibrating High Power Sensors

Post by csperrazza »

High Power Sensors, such as the Agilent 8481B and 8482B, have a dynamic range of 0 dBm to +37 dBm, (1 mW to 5 W), or higher. Using the practice of calibrating the middle of the sensors dynamic range (100 mW) can become costly because of the need for additional test equipment.

It has been proven that calibrating at 10 mW is typically high enough above the sensor's noise floor to be considered in the linear region. The specification for the Agilent 8481B (0 to 44 dBm) states a negligible linearity deviation up to +30 dBm, meaning the sensor’s response should be no different at 10 dBm than at 20 dBm.

The data provided below shows an 8481B calibrated at both 10mW and 100mW.
8481B (10mW).JPG
The System IIB is fully capable of measuring both the 8481B and 8482B power sensors as well as a wide variety of other sensors. The central component of the System IIB is the 1806A Dual Type IV Power Meter. This along with transfer standard such as the F1130A and SureCAL software, most sensors in the 100 kHz to 18 GHz range can be calibrated in less than 5 minutes.
system II.JPG
Some important things to consider when calibrating at 10 mW.
    There will be 6 dB of attenuation built into the F1130A (splitter) so your signal source will need to generate at least 16 dBm in order to calibrate at 10 mW. The addition of an amplifier such as the TEGAM 1727A may be needed to provide the necessary power across the desired frequency.
    You do not have the required permissions to view the files attached to this post.