I would like ask how should I configurate UUT monitor in SURECAL If I calibrate HP 478A opt H73 connected with1830A and F1130A connected with 1830A as STD?
Regards Peter
SureCAL Settings for HP478A H73 with 1830A as DUT monitor?
Re: SureCAL Settings for HP478A H73 with 1830A as DUT monito
Hi Peter,
To calibrate an HP 478 with an 1830A as a monitor do the following:
1. Select 1830A for monitor 2 in flex standards (This would be the 1830A connected to 478A)
2. The 1830A connected to the F1130A would still be monitor 1
3. In the procedure the UUT monitor would be changed to 1830A with 200 ohm resistance
That should do it.
Charlie
Feel free to contact me at my csperrazza@tegam.com or right back here on forums.
To calibrate an HP 478 with an 1830A as a monitor do the following:
1. Select 1830A for monitor 2 in flex standards (This would be the 1830A connected to 478A)
2. The 1830A connected to the F1130A would still be monitor 1
3. In the procedure the UUT monitor would be changed to 1830A with 200 ohm resistance
That should do it.
Charlie
Feel free to contact me at my csperrazza@tegam.com or right back here on forums.
Re: SureCAL Settings for HP478A H73 with 1830A as DUT monito
Hi,
Thank You for very quick answer. I had some doubts because "No compensation" mode was active when procedure has been running.
I thought "323 compatible" mode should be active. Procedure changed mode automatically to "no compenastion" when calibration has been started.
Peter
Thank You for very quick answer. I had some doubts because "No compensation" mode was active when procedure has been running.
I thought "323 compatible" mode should be active. Procedure changed mode automatically to "no compenastion" when calibration has been started.
Peter
Re: SureCAL Settings for HP478A H73 with 1830A as DUT monito
Peter,
It seems you have found a bug in SureCAL. I will work with the SureCAL team and get it resolved as soon as possibly.
Charlie
It seems you have found a bug in SureCAL. I will work with the SureCAL team and get it resolved as soon as possibly.
Charlie
Re: SureCAL Settings for HP478A H73 with 1830A as DUT monito
Issue Resolved:
If you edit the test profile and select 1830A for the UUT monitor a Vcomp check box to the right of the 1830A check box becomes available.
If you check the box the 1830A will be placed in 432 comp mode during the calibration.
If you edit the test profile and select 1830A for the UUT monitor a Vcomp check box to the right of the 1830A check box becomes available.
If you check the box the 1830A will be placed in 432 comp mode during the calibration.